| DSH 423
9 s s |5 D" w9 \ q1 K3 L3 }' `& _, s# t9 q
| Use of a PTC as a protective device' b/ K; d( f$ I; _- u
| 57.9 b)
1 Y H# V, n# ]2 X9 j/ z& I | 60601-1(ed.2);am1;am2
. C1 G2 U" g/ t3 V5 K) v$ u/ p |
! @( k! U u) k- S+ _Standard(s)- (year and edition):5 G" c# b0 ?* T9 q4 o! e9 [
IEC 60601-1:1988 Ed.2 Am1+Am22 e* N4 H2 k! }8 r8 |# V
Sub clause(s): 57.9 b) " `5 C3 W2 D T3 X. Q
Sheet n°: DSH-4234 m% V e# I" ]: \
Subject: Use of a PTC as a protective device
) A" [* {3 d3 P' r$ RKey words: PTC, protective device, Mains supply transformers, overload
) |: S- z7 f9 fConfirmed by CTL at its 39th meeting, in Cologne$ v! m1 Z6 h5 ^
Question:, n! P, T. l! B+ K& v2 l+ z, \, g
Is it permitted to use a PTC as protective device and what should be the overload test?
4 ^2 p2 S2 K/ i( }4 O YDecision: H5 N3 _% M$ W0 B- U
Where a PTC is used as a protective device for a mains supply transformer, the requirements for* h3 g* N0 x% U
THERMAL CUT-OUTS as required by sub-clause 57.9 shall be applied.: w5 C! s% A3 l; ^3 m6 J& W7 {
They shall be in accordance with International or National standards which address the reliability4 G, W) s1 L6 f$ m/ k- c* Q
of these components, e.g. as specified in IEC 60950 Amendment 4 clause 2.11 (Ed. 2).0 T$ i8 W# B- f
h% }& i" `, {% T7 G |