| Humidity test requirements) H1 [( l# `3 X- N$ k
| Various& f0 Q/ J% n* r r5 S9 o" F. E; o! S
| 60335-1(ed.4);am1;am2 & 61010-1(ed.2) & 60601-1(ed.3) & 60950-1(ed.2) & 60065(ed.7);am1
& ~; J# V' b# z3 t | 1 V; l* N8 W" _/ i
Question: Various IEC standards have requirements for humidity/temperature conditioning, but have different specifications for the test conditions.* Q3 w& L+ V; Q& [- j
IEC 60065 – 93 +2%/-3% (t 30°C +0/-2°C)
* Z9 n3 ~& j5 A" Z5 R4 VIEC 60950-1 – 91% - 95% (t 20-30°C +/-1°C)
& u% s; {5 o0 q5 |. a9 e! t) P7 f0 _IEC 60601-1 – 93 + 3%/-3% (t 20-30°C)0 F' P" k6 t0 T) |- o4 D* J
IEC 61010-1 – 92,5 + 2,5%/-2,5% (t 40°C +/- 2°C)
% X6 Q* V0 ^% mIEC 60335-1 – 93 + 3%/- 3% (t 20-30°C +/-1°C)
* t2 P' e) [, y9 T, CThere does not seem to be a technical justification for the differences in the specifications.& @; k' X" Z" D
Accordingly, it is suggested and recommended that the humidity/temperature specifications in the above standards be consistent.' ^0 t* Y9 P% d( o/ P2 Y/ U( ^2 q
Decision: The chamber is to be set at 93% RH, with +/-3% RH tolerance, at a convenient temperature between 20°C and 30°C, but the chosen temperature to be kept within +/- 2°C during the test.
. M! r( g( E* }2 ^) hThe standard specifications for the humidity are to be met at the reference/measuring point, where the controlling humidity sensor is located.. b; a2 E: o G$ {2 ]) w1 V* e
The tolerances permitted in CTL DSH 251A (+/-6% RH) are allowed in other parts of the humidity chamber, away from the reference point.8 p \+ x/ D0 b