| Preconditioning of equipment intended for conditioned rooms1 N* ?2 Y/ O( e. o
| 4.109 ~) L: }, l" O1 ^9 q* C% O! e
| 60601-1(ed.2);am1;am2 V1 o% g2 T" F% y4 F
| ' S) M; C# P% z
Standard(s)- (year and edition):4 ^( ^2 {% V3 `; Q: z# \0 V
IEC 60601-1:1988 Ed.2 Am1+Am2
# B, p' T& n4 z* Z7 q# CSub clause(s): 4.10* q* z6 _, P/ n
Sheet n°: DSH-4158 P1 R# n. c2 U4 B
Subject: Preconditioning of equipment intended for conditioned rooms
! t" t8 ?" D/ i, u: r' jKey words: Preconditioning/ i! t9 d& g% S; ^8 y0 O
Confirmed by CTL at its 39th meeting, in Cologne; Q' I7 \* j2 \' ^) c8 O1 q
Question:& \" ?" W7 E* m" ]3 ^
Should this test be applied to equipment which is permanently installed and operated only in a! o o& r7 R7 r+ M7 ^' s
controlled temperature and humidity conditions as specified in the ACCOMPANYING
8 \) m2 M6 W: ^7 ~6 `DOCUMENTS
- Q, M; _ j1 K4 B1 T& z$ lDecision:7 I; T- a2 X: T C3 J
No exception can be made from the requirements of Sub-clause 4.10. However, attention is
- K& O; A1 @* R/ \1 Hdrawn to the 3rd paragraph:
* P% R n& A2 u8 _; k; Y) WThe test shall be applied only to those EQUIPMENT parts likely to create a SAFETY HAZARD
3 `1 e; w- K! c, Z- }" ~when influenced by the climatic conditions that are simulated by the test.