In addition, double and reinforced insulation shall withstand the partial discharge test according to 7.5.3, if the recurring peak working voltage across the insulation is greater than 700 V and the voltage stress on the insulation is greater than 1 kV/mm.
) L% R& {( h, |! R/ _8 @. S. j, jCompliance is shown by impulse withstand voltage test according to 7.5.1 and the a.c. or d.c. voltage test according to 7.5.2, and partial discharge testing according to 7.5.3 if applicable.
; ~* \6 k- w9 H0 c" `4 T9 w3 E7 p& X2 R. U( e7 @$ i
Starting from a voltage below the rated discharge voltage UPD a, the voltage shall be linearly increased to 1,875 times UPD and held for a maximum time of 5 s) |5 G6 Z- N! X) ]. w
The voltage shall then be linearly decreased to 1,5 times UPD (± 5 %) and held for a maximum time of 15 s, during which the partial discharge is measured
7 s7 X$ J9 `1 l! a- I; O% WThe test shall be considered to have been successfully passed if the partial discharge is less than 10 pC during the measurement period1 A- l& ?9 K! {( P9 i