[LITE] PDSH 998 Measurement of circuit power factor and supply current

[复制链接]
跳转到指定楼层
楼主
发表于 2012-9-15 14:56 | 只看该作者 回帖奖励 |倒序浏览 |阅读模式
广东安规检测
有限公司提供:
PDSH 998
! `2 M4 _& H+ b. W) G9 r3 E

$ ~, O6 r9 K1 p+ C3 c' h
Measurement of circuit power factor and supply current
5 R/ Q) C2 ^4 C% Y2 S( }
General2 R7 q! S  ]( y& @# W) V6 ^
60968(ed.1);am1;am2 & 60969(ed.1);am1;am2
# r: ]) C9 |0 T% m1 Q

2 @) c! t5 Z5 q' ZStandard(s):
# U! H, W' H* }" p5 i. ~IEC 60968 All editions+ P7 N; e5 L3 _( r8 A2 w
IEC 60969 All editions- L' ^2 C* P- c; d7 t7 a. W7 p! r
Subclause(s): General$ V% h0 c6 x( l& U# t5 e& P
No. Year PDSH 0998 2010
6 v2 n* Q" e7 |- W4 Q/ M  oCategory: LITE
. i8 \: M8 G$ b! X9 L8 eDeveloped by: ETF5 OSM/LUM
. T8 I" H. M9 ESubject:
3 j- Q( B; P% R* l5 N, z/ t/ zMeasurement of circuit power factor and supply current
' ]6 D1 g4 o6 u0 Q5 B2 RKey words:# E5 e; m1 Y' w; k4 p" p+ R4 H0 H
- True power factor
1 q* K" t" F. ]6 ?- Supply current) `7 K4 X8 [* b. r/ S/ ?1 z1 r
- High frequency components
; `" O% R$ |- zTo be approved at the 49th CTL Plenary Meeting, in 20128 l0 L2 x- _" j  ^1 `) J; F
Question:2 K7 _( K- p$ ]8 N0 ]3 |
When measuring the circuit power factor and supply current of non linear load (a self-ballasted lamp), we
" T% l+ k+ s( Q$ O- Jwould like to check if there is a need to consider the frequency range of the power meter. The product in
* ^" N4 C2 J  \question produced high frequency components as shown in Figure 1. The true power factor can only be. a6 |- k: w! i; H
measured with power meter set to encompass these high frequencies. With the power meter’s frequency range& C) a4 G; P5 ]9 W
specified from d.c. to 25 KHz, the power meter reads a power factor of 0,880 and a current of 330 mA. As the
- Q$ K$ ?: w/ E3 u" T- bfrequency range was extended to 50 KHz, we found the power factor is much lower compared with cos f% }( |  n, x! N% E( {
(power factor measured at fundamental frequency). Which value of power factor can we accept on the product
1 k1 d2 D8 x0 v8 x0 L6 g+ }or on the manufacturer catalogue?) d0 v( d- J( h7 N7 a3 _5 u
Decision:& l) R9 l9 U0 B9 [% u; X9 F
In case of high harmonics and high crest factor, to get an accurate measurement, a power meter with a higher, U' Q, u% a4 k2 p
frequency range has to be used.
6 f" o* F' ?( M, R
- f" ], Y! z  r% c" `, C/ [) s% h7 x0 w% K& e0 P

1 w4 _) ^" K$ i) r# ^
& Z! C2 b( s" c# T' b& m7 l4 Y
: V0 w2 U% m  o- a. ^8 m
# N; ?# B6 Q! X$ T9 Z% A) ~1 J

本帖子中包含更多资源

您需要 登录 才可以下载或查看,没有帐号?注册安规

x
您需要登录后才可以回帖 登录 | 注册安规

本版积分规则

关闭

安规网为您推荐上一条 /2 下一条

QQ|关于安规|小黑屋|安规QQ群|Archiver|手机版|安规网 ( 粤ICP13023453-10 )

GMT+8, 2024-11-30 07:48 , Processed in 0.134643 second(s), 26 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.|广东安规赞助

快速回复 返回顶部 返回列表